  • User Non avatar #137100 United States10 months ago

    How do you switch from metic to standard?

  • User Non avatar #137135 Asia10 months ago

    1. Turn the monitor OFF (display blank).
    2. Press and hold the SET button for 3 seconds until the year begins flashing. 
    3. Keep pressing the SET button until 0.0 appears in the display (unit-selection mode).
    4. Press the MEM button 1 time, and it'll change from “0.0” to “0”, switching from kPa to mmHg. 
    5. Press the SET button, the display shows “donE” and the device turns off.

  • How do you switch the monitor back to mmHg from kPa.

  • Need to switch back to standard from metic. HOW?

  • User Non avatar #137099 United States10 months ago

    Need to switch from metic to standard

  • User Non avatar #3911 United States5 years ago

    It won't stop inflating and won't stop when pushing the stop button

  • User Non avatarThomas W Z5 years ago

    I keep getting an E20 error message. What's up?

  • User Non avatar #137115 Asia10 months ago

    E20 shows means the measurement process does not detect the pulse signal. Please try to loosen the clothing on the arm and then measure again.